Stubbie wrote:The color of the wires coming off are light blue and red.
That matches a type T. The color code for a type J is different.
If you do actually have a type T, and the PID controller thinks it is a type J, then the readings will be off by a mile.
Example: if the real boiler temp is 220F, and the ambient temp at the PID is 80F, then the measured voltage from a type T would be around 3.427 mV. If the PID thinks the sensor is a type J, however, it will still read the same 3.427 mV, but will conclude that the temperature of the boiler is 104F. That is a 116F reading error, in the direction that overheats the boiler.
Jim